Cabreira et al: Development and Finite Element Analysis of a Patient-Specific Implant for Atlantoaxial Joint Stabilization via Dorsal Approach in Dogs
Veterinary and Comparative Orthopaedics and Traumatology 1, 2026

Development and Finite Element Analysis of a Patient-Specific Implant for Atlantoaxial Joint Stabilization via Dorsal Approach in Dogs

🔍 Key Findings

  • A patient-specific dorsal implant for atlantoaxial stabilization was developed using CT-based 3D modeling, enabling customized implant design and safe screw trajectory planning.
  • The implant accommodates six 1.7-mm bicortical locking screws, including two transarticular screws plus two screws in each vertebra (atlas lateral masses and axis lamina/spinous base).
  • The implant also functions as a drilling guide, allowing the screw-head housing to accept a threaded drill guide that ensures accurate drilling along predefined safe corridors.
  • An opening around the C1–C2 intervertebral foramen was incorporated to avoid injury to the C2 nerve root during dorsal instrumentation.
  • Finite element analysis evaluated implant behavior under flexion, extension, lateral flexion, and torsion using loads of 50 N (physiologic) and 107 N (supraphysiologic).
  • Maximum stress occurred during lateral flexion with axis fixation (425 MPa), which remained well below the titanium alloy yield strength of 880 MPa, indicating no mechanical failure.
  • Maximum implant displacement was minimal (0.13 mm), far less than the approximately 2 mm movement of the intact atlantoaxial joint under physiologic loading.
  • The implant design demonstrated adequate rigidity and mechanical safety in simulation, suggesting potential to improve dorsal stabilization while avoiding complications associated with PMMA constructs.

Simini Surgery Review Podcast

How critical is this paper for crushing the Boards?

🚨 Must-know. I’d bet on seeing this.

📚 Useful background, not must-know.

💤 Skip it. Doubt it’ll ever show up.

Thanks for the feedback!
We'll keep fine-tuning the articles vault.
Oops — didn’t go through.
Mind trying that again?

Cabreira et al: Development and Finite Element Analysis of a Patient-Specific Implant for Atlantoaxial Joint Stabilization via Dorsal Approach in Dogs
Veterinary and Comparative Orthopaedics and Traumatology 1, 2026

🔍 Key Findings

  • A patient-specific dorsal implant for atlantoaxial stabilization was developed using CT-based 3D modeling, enabling customized implant design and safe screw trajectory planning.
  • The implant accommodates six 1.7-mm bicortical locking screws, including two transarticular screws plus two screws in each vertebra (atlas lateral masses and axis lamina/spinous base).
  • The implant also functions as a drilling guide, allowing the screw-head housing to accept a threaded drill guide that ensures accurate drilling along predefined safe corridors.
  • An opening around the C1–C2 intervertebral foramen was incorporated to avoid injury to the C2 nerve root during dorsal instrumentation.
  • Finite element analysis evaluated implant behavior under flexion, extension, lateral flexion, and torsion using loads of 50 N (physiologic) and 107 N (supraphysiologic).
  • Maximum stress occurred during lateral flexion with axis fixation (425 MPa), which remained well below the titanium alloy yield strength of 880 MPa, indicating no mechanical failure.
  • Maximum implant displacement was minimal (0.13 mm), far less than the approximately 2 mm movement of the intact atlantoaxial joint under physiologic loading.
  • The implant design demonstrated adequate rigidity and mechanical safety in simulation, suggesting potential to improve dorsal stabilization while avoiding complications associated with PMMA constructs.

Simini Surgery Review Podcast

Know What Matters in the Literature - and Why

We distill peer-reviewed surgical studies into clinically relevant summaries and
exam-style questions, so you can make informed decisions with confidence.

Free Access. No Spam. Just Smarter Surgical Learning

Multiple Choice Questions on this study

In Cabreira 2026 et al., on dorsal AAI implant, what was the maximum stress recorded in the implant during finite element testing?

A. 120 MPa
B. 250 MPa
C. 425 MPa
D. 880 MPa
E. 1200 MPa

Answer: 425 MPa

Explanation: The peak stress was 425 MPa, which remained below the titanium alloy yield strength of 880 MPa, indicating no predicted failure.
In Cabreira 2026 et al., on dorsal AAI implant, how many screws were incorporated into the patient-specific implant design for atlantoaxial stabilization?

A. Four screws total placed only in the atlas
B. Five screws including one transarticular screw
C. Six bicortical locking screws including two transarticular screws
D. Eight screws including bilateral ventral lag screws
E. Ten screws spanning C1 to C3

Answer: Six bicortical locking screws including two transarticular screws

Explanation: The implant design specifically included six 1.7-mm bicortical locking screws: two transarticular screws plus two screws each in the atlas and axis.
In Cabreira 2026 et al., on dorsal AAI implant, which loading condition produced the highest stress during finite element analysis?

A. Flexion with atlas fixation
B. Extension with axis fixation
C. Lateral flexion with axis fixation
D. Torsion with atlas fixation
E. Axial compression

Answer: Lateral flexion with axis fixation

Explanation: The highest von Mises stress occurred during lateral flexion with the axis fixed under a 107 N load.
In Cabreira 2026 et al., on dorsal AAI implant, what was the maximum implant displacement observed during finite element analysis?

A. 0.13 mm
B. 0.5 mm
C. 1 mm
D. 2 mm
E. 4 mm

Answer: 0.13 mm

Explanation: The largest displacement recorded was approximately 0.13 mm under a 107 N load, indicating a highly rigid construct compared with physiologic joint movement.
In Cabreira 2026 et al., on dorsal AAI implant, what additional surgical function does the implant provide during instrumentation?

A. It serves as an external reduction clamp
B. It functions as a polymethylmethacrylate mold
C. It acts as a drill guide for screw placement
D. It provides ventral decompression of the spinal cord
E. It measures intraoperative joint stability

Answer: It acts as a drill guide for screw placement

Explanation: The screw-head housing accepts a threaded drill guide so the implant itself acts as a template for drilling safe screw corridors.

Access the full library of surgical summaries and exam-style questions.

Educational content developed independently and supported by Simini.

The maker of Simini Protect Lavage.

Something off with this article?
Tell us what needs fixing—drop your note below.
Thanks for flagging this!
We’ll review your note as soon as possible.
Oops! Something went wrong while submitting the form.
Anything to add to this article?
Know a finding or detail we missed? Share it below.
Thanks for the suggestion!
We’ll review it as soon as possible.
Oops! Something went wrong while submitting the form.