Your Custom Quiz

In Lopez Barroso 2026 et al., on sacroiliac screw accuracy, what was the primary method used to confirm optimal drill trajectory?

Thank you! Your submission has been received!
Oops! Something went wrong while submitting the form.
Correct. The device used a stainless-steel sphere and crosshair K-wires to confirm fluoroscopic alignment.
Incorrect. The correct answer is Superimposition of radiopaque crosshairs and sphere.
The device used a stainless-steel sphere and crosshair K-wires to confirm fluoroscopic alignment.

🔍 Key Findings

  • Mean SI joint reduction was 94.25%, exceeding the 90% target considered important to reduce screw loosening
  • Mean sacral width purchase was 82.52%, with all screws achieving >60%—a threshold for reducing implant loosening risk
  • All 20 screws were accurately placed within the sacral body, confirmed via CT or radiographs
  • A 3D-printed drill guide and table-bound system facilitated consistent screw placement via minimally invasive technique
  • Craniocaudal and dorsoventral angles were well controlled (mean CCA = –1.22°, DVA = 0.71°), demonstrating accurate trajectory
  • No intraoperative complications were reported; all pilot holes drilled on first attempt
  • Use of fluoroscopy and orthogonal imaging enhanced surgical accuracy and reduced malposition risk
  • The system permitted screw placement through a sleeve that doubled as a muscle retractor, allowing implant delivery without disruption

Lopez Barroso

Veterinary Surgery

1

2026

Accuracy of a table‐bound reduction and drill guide system for minimally invasive sacroiliac screw placement in cats

2026-1-VS-lopezbarroso-3

Article Title: Accuracy of a table‐bound reduction and drill guide system for minimally invasive sacroiliac screw placement in cats

Journal: Veterinary Surgery

In Lemmon 2025 et al., on synovitis severity scoring, what was the most common synovitis severity score observed during arthroscopy?

Thank you! Your submission has been received!
Oops! Something went wrong while submitting the form.
Correct. Score 3/5 was the most frequently observed severity grade, present in 55.2% of joints.
Incorrect. The correct answer is 3/5.
Score 3/5 was the most frequently observed severity grade, present in 55.2% of joints.

🔍 Key Findings

Synovitis was present in 100% of canine stifles with CCL disease (n = 163).

The most frequent severity score was 3/5 (55.2%), followed by 4/5 (24.5%).

Higher synovitis scores were significantly associated with:

  • Higher median cartilage scores (p = .042, OR = 2.1 per unit increase)
  • Longer duration of clinical signs (p < .001, OR = 1.27 per month)

Bodyweight (p = .083) and sex (p = .17) were not statistically significant in multivariable analysis.

Bucket handle meniscal tears were not associated with synovitis severity.

Clinical implication: Earlier intervention may help reduce synovitis and slow OA progression.

Lemmon

Veterinary Surgery

3

2025

Arthroscopic synovitis severity scoring in canine stifles with cranial cruciate ligament disease

2025-3-VS-lemmon-1

Article Title: Arthroscopic synovitis severity scoring in canine stifles with cranial cruciate ligament disease

Journal: Veterinary Surgery

In Cormillot 2026 et al., on laparoscopic handle ergonomics, surgeons with smaller hands most commonly preferred which instrument characteristic?

Thank you! Your submission has been received!
Oops! Something went wrong while submitting the form.
Correct. Smaller-handed surgeons preferred reduced loop spacing and alternative finger placement patterns to improve comfort and control.
Incorrect. The correct answer is Smaller finger-loop spacing.
Smaller-handed surgeons preferred reduced loop spacing and alternative finger placement patterns to improve comfort and control.

🔍 Key Findings

  • Surgeon hand size significantly influences preferred finger-loop handle configuration when using laparoscopic instruments.
  • Smaller-handed surgeons preferred smaller loop spacing and different finger placements, suggesting standard instrument designs may not accommodate all users.
  • Larger-handed surgeons favored wider finger-loop spacing, which improved comfort and maneuverability during laparoscopic manipulation.
  • Instrument ergonomics varied substantially between users, demonstrating that a “one-size-fits-all” laparoscopic handle design is suboptimal.
  • Hand anthropometry affected both perceived comfort and functional performance during simulated laparoscopic tasks.
  • Loop positioning and finger insertion pattern influenced wrist posture and grip mechanics, potentially affecting fatigue during prolonged procedures.
  • Ergonomic mismatch between surgeon hand size and handle design may contribute to surgeon discomfort and reduced efficiency during minimally invasive surgery.
  • The study supports development of adjustable or size-specific laparoscopic handles to improve surgeon ergonomics and operative performance.

Cormillot

Veterinary Surgery

2

2026

Surgeon hand size influences laparoscopic finger loop handle preference

2026-2-vs-cormillot-2

Article Title: Surgeon hand size influences laparoscopic finger loop handle preference

Journal: Veterinary Surgery

In Barrett 2026 et al., on scapula plating biomechanics, which outcome was significantly improved with double plate fixation compared to single plate fixation?

Thank you! Your submission has been received!
Oops! Something went wrong while submitting the form.
Correct. Double plating significantly increased stiffness (563.7 vs 392.8 N/mm; p < .0001), indicating a stronger construct.
Incorrect. The correct answer is Increased stiffness.
Double plating significantly increased stiffness (563.7 vs 392.8 N/mm; p < .0001), indicating a stronger construct.

🔍 Key Findings

  • Double plate fixation significantly increased construct stiffness (563.7 vs 392.8 N/mm; p < .0001), supporting superior biomechanical stability
  • Displacement was significantly reduced with double plating (−0.48 mm vs −0.81 mm; p < .0001), indicating improved resistance to cyclic loading
  • No difference in primary plate strain between single and double constructs, suggesting load-sharing did not reduce plate stress
  • Caudolateral placement of the second plate enhances biomechanical advantage by increasing moment arm and resisting tensile forces
  • Cyclic loading model (7200 cycles) simulated 1 hour of walking, making findings clinically relevant to early postoperative activity
  • Strain across fracture gap was lower in double plate constructs (~4.3% vs ~6.9%), supporting improved conditions for fracture healing
  • No construct failures occurred, but actuator slippage occurred only in double plate group, highlighting potential testing/artifact considerations
  • Clinical implication: double plating may allow earlier weightbearing and reduce need for external coaptation, improving recovery outcomes

Barrett

Veterinary Surgery

3

2026

Double plate fixation improves stiffness in a comminuted canine scapula fracture gap model compared to single plate fixation

2026-3-vs-barrett-1

Article Title: Double plate fixation improves stiffness in a comminuted canine scapula fracture gap model compared to single plate fixation

Journal: Veterinary Surgery

In Davies 2024 et al., on lymphaticovenous anastomosis, what was used to create the end-to-end connection?

Thank you! Your submission has been received!
Oops! Something went wrong while submitting the form.
Correct. A commercially available MAC device enabled end-to-end coupling of TD to AV.
Incorrect. The correct answer is Microvascular anastomotic coupler (MAC).
A commercially available MAC device enabled end-to-end coupling of TD to AV.

🔍 Key Findings

  • Lymphaticovenous anastomosis (TD to AV) was successfully performed in all 8 feline cadavers using a microvascular anastomotic coupler (MAC).
  • Anastomotic patency was confirmed intraoperatively and postoperatively in 7/8 cats via contrast lymphography or retrograde venography.
  • Dissection and anastomosis took a median of 120 minutes, with minimal technical complications.
  • A 1.5 mm MAC was used in 6 cats, and a 2.0 mm in 2 cats; TD diameter ranged 1.0–1.5 mm, AV up to 2.25 mm.
  • Challenges included vessel twisting and luminal patency issues, resolved intraoperatively with minor adjustments (e.g., repeat pinning, tacking suture).
  • MAC use eliminated need for hand-suturing, lowering skill demands but requiring precise alignment.
  • Technique provides direct lymphatic-to-venous drainage, potentially reducing the stimulus for collateral vessel formation.
  • May serve as a future treatment option for feline idiopathic chylothorax, warranting further in vivo studies.

Davies

Veterinary Surgery

7

2024

Lymphaticovenous anastomosis of the caudal thoracic duct to the azygous vein: A feline cadaver study

2024-7-VS-davies-1

Article Title: Lymphaticovenous anastomosis of the caudal thoracic duct to the azygous vein: A feline cadaver study

Journal: Veterinary Surgery

In Walter de Bruyn 2024 et al., what mechanical property was completely normalized across all working lengths after orthogonal plate application?

Thank you! Your submission has been received!
Oops! Something went wrong while submitting the form.
Correct. Addition of orthogonal plates eliminated differences in bending stiffness across all working lengths:contentReference[oaicite:3]{index=3}
Incorrect. The correct answer is Bending stiffness.
Addition of orthogonal plates eliminated differences in bending stiffness across all working lengths:contentReference[oaicite:3]{index=3}

🔍 Key Findings Summary

  • Primary 3.5-mm LCP used with short (SWL), medium (MWL), and long (LWL) working lengths
  • Addition of orthogonal 2.7-mm LCP resulted in:
    • Significantly higher bending stiffness for SWL, MWL, and LWL (p < 0.0001)
    • Higher torsional stiffness for MWL and LWL (not for SWL)
    • Significantly lower strain across all working lengths in bending (p < 0.01)
  • Working length inversely related to construct stiffness and directly to plate strain
  • Orthogonal plates eliminated stiffness differences across working lengths in bending
  • Suggests orthogonal plates can improve implant fatigue life and allow compensation when short working lengths are unachievable

Walterdebruyn

Veterinary and Comparative Orthopedics and Traumatology

4

2024

Effect of an Orthogonal Locking Plate and Primary Plate Working Length on Construct Stiffness and Plate Strain in an In vitro Fracture-Gap Model

2024-4-VCOT-walterdebruyn-4

Article Title: Effect of an Orthogonal Locking Plate and Primary Plate Working Length on Construct Stiffness and Plate Strain in an In vitro Fracture-Gap Model

Journal: Veterinary and Comparative Orthopedics and Traumatology

In Dekerle 2022 et al., on ectopic ureter correction, which surgical technique was associated with the **lowest rate of minor complications**?

Thank you! Your submission has been received!
Oops! Something went wrong while submitting the form.
Correct. CLA had a 13% minor complication rate compared to 100% with neoureterostomy.
Incorrect. The correct answer is Cystoscopic-guided laser ablation (CLA).
CLA had a 13% minor complication rate compared to 100% with neoureterostomy.

🔍 Key Findings

  • Cystoscopic-guided laser ablation (CLA) was associated with significantly fewer minor complications (13%) than neoureterostomy (100%) (P < .01)
  • CLA resulted in significantly fewer recurrences of incontinence compared to neoureterostomy (0/7 vs 5/12; P < .05)
  • 80% of dogs achieved continence within 1 month postoperatively, with a median continence score of 10
  • Long-term continence was achieved in 88% of dogs, with or without adjunctive treatment, over a median of 66 months
  • Major complications occurred in only 8% of dogs, and all were successfully managed with surgical revision
  • Persistent ureteral remnants were seen only in neoureterostomy dogs, potentially contributing to incontinence recurrence
  • All dogs with incontinence recurrence responded to medical treatment, while only 1/5 with persistent incontinence after surgery did
  • CLA is recommended over open surgery for iEU correction, due to lower complication and recurrence rates

Dekerle

Veterinary Surgery

4

2022

Outcomes of 25 female dogs treated for ectopic ureters by open surgery or cystoscopic-guided laser ablation

2022-4-VS-dekerle-1

Article Title: Outcomes of 25 female dogs treated for ectopic ureters by open surgery or cystoscopic-guided laser ablation

Journal: Veterinary Surgery

In Story 2024 et al., on surgical correction of excessive tibial plateau angle (eTPA), which two techniques demonstrated the most accurate achievement of their target TPA values?

Thank you! Your submission has been received!
Oops! Something went wrong while submitting the form.
Correct. Groups A and D had the least variation from their target TPAs, with mean correction accuracy near 1.0.
Incorrect. The correct answer is CBLO + CCWO and PTNWO.
Groups A and D had the least variation from their target TPAs, with mean correction accuracy near 1.0.

🔍 Key Findings

  • All four techniques achieved TPA <14°, meeting the threshold for acceptable surgical correction in eTPA cases.
  • Group A (CBLO + CCWO) and Group D (PTNWO) showed highest accuracy in achieving target TPA values.
  • Group B (TPLO + CCWO) resulted in significant tibial shortening compared to other techniques.
  • Group A caused the greatest cranial mechanical axis shift, while Group B caused the least.
  • Group C (mCCWO) resulted in consistent under-correction of TPA, despite aiming for 0°.
  • Modified or neutral wedge osteotomies (Groups C and D) had minimal effect on tibial length, making them suitable when preservation is important.
  • All techniques involved mechanical axis shifts, highlighting the importance of preoperative planning to minimize morphologic disruption.
  • Supplemental fixation was standard for all procedures to reduce risks such as tibial tuberosity fracture and plateau leveling loss.

Story

Veterinary Surgery

8

2024

Morphologic impact of four surgical techniques to correct excessive tibial plateau angle in dogs: A theoretical radiographic analysis

2024-8-VS-story-3

Article Title: Morphologic impact of four surgical techniques to correct excessive tibial plateau angle in dogs: A theoretical radiographic analysis

Journal: Veterinary Surgery

In Drudi 2022 et al., on CAL vs TAL outcomes, what clinical difference was reported between the two groups at 15-day follow-up?

Thank you! Your submission has been received!
Oops! Something went wrong while submitting the form.
Correct. Both groups had improved clinical signs with no differences at day 15.
Incorrect. The correct answer is No clinical difference was observed.
Both groups had improved clinical signs with no differences at day 15.

🔍 Key Findings

  • Cricoarytenoid lateralization (CAL) resulted in a significantly greater increase in rima glottidis area at both immediate (205%) and 15-day (199%) time points compared to thyroarytenoid lateralization (TAL) (152% and 127%, respectively).
  • TAL group showed a significant reduction in rima glottidis area between immediate and 15-day postoperative measurements (P < .05), while CAL group had no significant reduction over time.
  • No dogs in either group showed postoperative complications, including aspiration pneumonia, at the 15-day follow-up.
  • All dogs showed improved clinical signs, including decreased stridor and increased exercise tolerance by day 15.
  • CAL produced more stable postoperative glottic area, potentially due to preserved anatomical tension, whereas TAL might experience tension loss due to thyroid cartilage repositioning.
  • Both procedures were technically effective and performed under the same protocol by a single board-certified surgeon.
  • Endoscopic image analysis was used to quantify rima glottidis area, demonstrating a reliable objective method for surgical outcome assessment.
  • Clinical outcome did not differ between groups, despite CAL showing a larger rima glottidis area.

Drudi

Veterinary Surgery

3

2022

Comparison of immediate and short‐term outcomes of cricoarytenoid and thyroarytenoid lateralization in dogs with idiopathic laryngeal paralysis

2022-3-VS-drudi-4

Article Title: Comparison of immediate and short‐term outcomes of cricoarytenoid and thyroarytenoid lateralization in dogs with idiopathic laryngeal paralysis

Journal: Veterinary Surgery

In Muroi 2025 et al., on refracture risk, what screw-to-bone diameter ratio (SBDR) is suggested as a **threshold** for increased refracture risk in growing dogs?

Thank you! Your submission has been received!
Oops! Something went wrong while submitting the form.
Correct. An SBDR over 0.4 may weaken bone, especially when screws become relatively large due to growth.
Incorrect. The correct answer is >0.4.
An SBDR over 0.4 may weaken bone, especially when screws become relatively large due to growth.

🔍 Key Findings

  • Refracture occurred in 5.5% of limbs, with higher incidence in the plate removal group (12.5%) vs. non-removal (3.5%).
  • In the non-plate removal group, refractures occurred at the most distal screw site, linked to greater screw position change during growth (OR 1.79, p=0.04).
  • Screw-to-bone diameter ratio (SBDR) >0.4 was a significant risk factor for refracture in the plate retention group.
  • In the plate removal group, refractures occurred at the original fracture site, associated with lower pixel value ratio (bone mineral density) and reduced radial thickness.
  • Implant-induced osteoporosis (IIO) beneath the plate likely contributed to refracture risk after plate removal.
  • Younger age at fracture (<6 months) was associated with higher refracture risk due to ongoing radial growth and shifting screw position.
  • No significant association was found between refracture and plate type (locking vs conventional), fixation method, or ulnar union.
  • Recommendations include careful SBDR sizing, motion restriction, and cautious plate removal decisions in growing dogs.

Muroi

Veterinary and Comparative Orthopaedics and Traumatology

2

2025

A Retrospective Study of Risk Factors Associated with Refracture after Repair of Radial–Ulnar Fractures in Small-Breed Dogs

2025-2-VCOT-muroi-2

Article Title: A Retrospective Study of Risk Factors Associated with Refracture after Repair of Radial–Ulnar Fractures in Small-Breed Dogs

Journal: Veterinary and Comparative Orthopaedics and Traumatology

Quiz Results

Previously Missed Questions
70%

You answered 7 out of 10 questions correctly

Question 1:

❌ Incorrect. You answered: Answer

Correct answer:

Rationale

Question 1:

✅ Correct! You answered: Answer

Rationale

Author: Journal Name - 2025

Article Title

Key Findings

Something off with this question?
Tell us what needs fixing—drop your note below.

You’re flagging: [question text]

Thanks for your feedback!
We’ll review your comment as soon as possible.
Oops! Something went wrong while submitting the form.