In Folk 2025 et al., on vessel sealing device reuse, what type of sterilization method was used after splenectomy?
🔍 Key Findings
40 dogs underwent splenectomy using 16 bipolar vessel sealing devices (VSDs)
Devices were reused up to 4 times after handwashing and ethylene oxide (EtO) sterilization
Biologic debris was found in 100% of devices, specifically under the transection blade, even after a single use
- Mostly scant (14/16) or mild (2/16) debris
No devices or debris yielded positive aerobic culture after EtO sterilization
EtO sterilization proved microbiologically effective despite visible residue
Perioperative failure rate: 1 device (malfunctioned during first activation)
Veterinary Surgery
3
2025
Incidence of residual biologic debris and contamination of reused bipolar vessel sealing devices after ethylene oxide sterilization following splenectomy
2025-3-VS-folk-1
In De Moya 2025 et al., on antebrachial deformity correction, what was the most common radiographic outcome following CESF and distraction osteogenesis?
🔍 Key Findings
- CESF with distraction osteogenesis restored elbow congruity and normalized aLDRA in skeletally immature dogs with PCDRP.
- Radial head subluxation was eliminated in all dogs, and elbow incongruity reduced significantly (from 6.1 mm to 0.3 mm, p <.01).
- Mean radial lengthening of 22.6 mm (∼11% of normal length) was achieved, but only 80% of recorded distraction translated to length gain.
- Major complications occurred in 2/12 dogs: one with permanent carpal contracture, one with radial fracture at wire tract.
- Minor complications (e.g., carpal pain, restricted extension, synostosis, pin tract issues) were noted in 10/12 dogs but generally resolved.
- Owner surveys (8/12 dogs) reported good to excellent long-term function, even up to 6 years post-op.
- Radial valgus deformities were moderate (mean 15°) and less severe than deformities from ulnar physeal closure.
- Surgical strategy included staged distraction, with radial or combined radius/ulna distraction guided by fluoroscopy and adjusted per case.
Veterinary Surgery
6
2025
Treatment of antebrachial deformities secondary to premature closure of the distal radial physis using circular external skeletal fixation and distraction osteogenesis in skeletally immature dogs
2025-6-VS-demoya-1
In Evers 2022 et al., on bone-to-tendon plate fixation, what was the primary reason for anchoring the plate to the tendon instead of the bone?
🔍 Key Findings
- Bone-to-tendon plate fixation allowed successful stabilization of a highly comminuted calcaneus fracture in a dog with fragments too small for traditional fixation.
- The plate was sutured to the common calcaneal tendon using a figure-of-8 pattern, bypassing the need for screw fixation into small proximal fragments.
- Radiographic union was achieved by 17 weeks, though considered delayed, with the dog returning to normal function by 36 weeks post-op.
- A second surgery was required to replace the tendon-anchored plate with a calcaneus-only plate due to skin ulceration and implant prominence.
- Implant-associated infection was suspected; cultures confirmed Staphylococcus pseudintermedius, managed with doxycycline and clindamycin.
- Use of human placental matrix (hPM) and both autogenous and allogenic bone grafts supported healing, though their specific contribution remains uncertain.
- Postoperative complications included delayed union and skin ulceration, emphasizing challenges of implant design and soft tissue management.
- This is the first report of using a bone-to-tendon plate for a calcaneus fracture in dogs and demonstrates its potential in cases where traditional methods are not viable.
Veterinary Surgery
5
2022
Use of a bone‐to‐tendon plate to stabilize a comminuted calcaneus fracture in a dog
2022-5-VS-evers-1
In Meltzer 2022 et al., on femoral implant selection, what was the most common major complication reported?
🔍 Key Findings
- Hybrid implants were used in older, heavier dogs with lower canal flare index (CFI) compared to cementless types.
- CFI <1.8 was associated with higher risk of femoral fracture or stem subsidence with cementless implants.
- Total complication rate was 14%, with catastrophic complications in only 1.5%—lower than reported in other studies.
- No significant difference in complication rates across implant types (BFX, BFX-C, hybrid).
- Femur fractures were rare (2.9%), lower than in prior studies, and successfully managed with cerclage/plate fixation.
- Coxofemoral luxation was the most common complication (8 cases); all occurred within 62 days post-op, mostly resolved with revision.
- Collared BFX stems did not significantly reduce complications, though they may limit stem subsidence.
- An implant selection algorithm was proposed, using age ≥7 years, weight ≥45 kg, and CFI <1.8 to guide cemented vs. cementless stem use.
Veterinary Surgery
2
2022
Case factors for selection of femoral component type in canine hip arthroplasty using a modular system
2022-2-VS-meltzer-3
In Poggi 2025 et al., on laparoscopic cholecystectomy in cats, what was the most commonly reported indication for surgery?
🔍 Key Findings
- Laparoscopic cholecystectomy (LC) was successfully performed in all 22 cats without conversion to open surgery.
- Complication rate was low: 3 cats had postoperative complications (1 seroma, 1 vomiting, 1 EHBDO and death).
- Median operating time was 41 minutes; median hospitalization was 3 days.
- No cases required conversion to open laparotomy, even in cats as small as 2.5 kg.
- Common devices used: Hemoclips or Hem-o-lok for cystic duct ligation; Ligasure, harmonic scalpel, or j-hook for dissection.
- Most common indications: cholelithiasis (9/22) and cholecystitis (10/22), with one biliary mucocele and one adenoma.
- Histopathology confirmed diagnosis in all cases; concurrent liver or intestinal biopsies were performed in some cats.
- Postoperative AUS and bloodwork were routine and important for detecting early signs of EHBDO.
Veterinary Surgery
5
2025
Laparoscopic cholecystectomy in 22 cats (2018–2024)
2025-5-VS-poggi-1
In Scheuermann 2023 et al., on MIPO with 3D-printed bone models, what was the main advantage of using the fracture reduction system (FRS) compared to intramedullary pin (IMP) reduction?
🔍 Key Findings
- Precontoured plates based on 3D-printed femurs produced accurate femoral alignment (median deviations <3 mm or <3° in all planes).
- Both fracture reduction system (FRS) and intramedullary pin (IMP) methods achieved near-anatomic alignment in cadaveric femoral fractures.
- FRS required fewer fluoroscopic images (median 7 vs. 26, P = .001), but longer surgical time (median 43 vs. 29 min, P = .011).
- Sagittal plane alignment: FRS led to mild increased recurvatum (median 2.9°), but still within near-anatomic limits (<5°).
- Axial alignment: Both groups achieved near-anatomic torsion (<10°), though one IMP case had acceptable (not near-anatomic) alignment.
- Custom drill guides and FRS improved fluoroscopy efficiency but were cumbersome and time-consuming to use. Authors do not recommend current prototype for clinical use.
- Clinical significance: 3D printed models allow accurate precontouring, reducing intra-op plate adjustment; custom guides may reduce radiation exposure for the surgical team.
Veterinary Surgery
7
2023
Minimally invasive plate osteosynthesis of femoral fractures with 3D-printed bone models and custom surgical guides: A cadaveric study in dogs
2023-7-VS-scheuermann-1
In Rocheleau 2023 et al., on shoulder stabilization, what was the primary advantage of the suture-toggle technique compared to bone anchors?
🔍 Key Findings
- Arthroscopically assisted stabilization with an IAD was feasible in cadaveric canine shoulders using both bone anchor and suture-toggle techniques.
- Suture-toggle repair was successfully performed in all specimens, while bone anchors could not be deployed in 3/10 shoulders, indicating lower feasibility.
- Median surgical time was shorter for the suture-toggle group (25.5 min) compared to the anchor group (37.3 min).
- Postoperative abduction angles returned to baseline in the anchor group, while the suture-toggle group had slightly reduced angles, suggesting possible overtightening.
- Violation of the articular surface occurred in 2 anchor specimens and 1 suture-toggle specimen, but none were considered likely to be clinically significant.
- Overall targeting accuracy using the IAD was 88%, supporting its utility in guiding tunnel/anchor placement.
- CT measurements showed acceptable bone stock and insertion angles, but anchor insertion angles were lower than optimal for mechanical pullout strength.
- Suture-toggle technique was considered simpler and more consistent, with fewer complications and faster execution than anchor placement.
Veterinary Surgery
4
2023
Ex vivo evaluation of arthroscopically assisted shoulder stabilization in dogs using an intra-articular aiming device
2023-4-VS-rocheleau-1
In Israel 2023 et al., on povidone-iodine lavage, which component best describes the composition and method of PrePIL?
🔍 Key Findings
- No infections (0/102) occurred in cases where preclosure povidone-iodine lavage (PrePIL) was used, compared to 21/2111 (0.99%) with saline lavage.
- Break-even cost for PrePIL was calculated at $49.74, while the actual cost was only $2.07, confirming strong cost-effectiveness.
- No adverse reactions or healing complications were reported in the 102 PrePIL cases.
- The PrePIL protocol used 0.35% povidone-iodine for 3 minutes before closure, followed by sterile saline lavage.
- Majority of THRs (97%) used cementless fixation (BFX) in PrePIL group; only 3 cases used CFX components.
- Surgeries followed consistent perioperative antibiotic protocols, making PrePIL the only major protocol variable.
- Antibiotics were not added to lavage, aligning with WHO and CDC guidelines for antimicrobial stewardship.
- The infection reduction with PrePIL supports routine use in total hip replacement to reduce periprosthetic joint infections.
Veterinary Surgery
1
2023
Preclosure povidone‐iodine lavage in total hip replacement surgery: Infection outcomes and cost–benefit analysis
2023-1-VS-israel-3
In Hertel 2025 et al., on portal venotomy for insulinoma, what was the primary reason the authors selected portal venotomy instead of en bloc vessel resection?
🔍 Key Findings
- Portal venotomy enabled successful en bloc resection of a pancreatic insulinoma and associated thrombus in the portal vein, a novel approach in veterinary literature.
- Temporary portal vein occlusion (15 min) was well tolerated, causing only transient small intestinal congestion without lasting hemodynamic compromise.
- No intraoperative hemorrhage occurred following vascular isolation and venotomy repair with 5–0 polypropylene in a simple continuous pattern.
- Postoperative complications were minimal, with only moderate pancreatitis managed medically and no portal hypertension observed on follow-up imaging.
- Histopathology confirmed insulinoma, with no initial metastasis; CT at 6 months revealed hepatic and peritoneal metastases but no local recurrence at the venotomy site.
- Portal vein luminal stenosis occurred post-repair but was clinically insignificant.
- Advanced CT imaging and 3D printed modeling significantly enhanced surgical planning and anatomical assessment.
- Surgical excision achieved good short-term outcomes, and the dog was alive and euglycemic one year later despite metastatic disease.
Veterinary Surgery
5
2025
Successful venotomy for portal tumor thrombus removal due to pancreatic carcinoma in a dog
2025-5-VS-hertel-1
In Araos 2024 et al., on helmet CPAP in brachycephalic postoperative dogs, what was the helmet CPAP treatment protocol in terms of pressure and duration?
🔍 Key Findings
- CPAP at 5 cmH₂O via helmet significantly improved PaO₂/FiO₂ ratio in the early postoperative period of brachycephalic dogs, supporting enhanced pulmonary oxygen exchange (p = .04).
- CPAP had no significant effect on SpO₂ (p = .9), PaCO₂ (p = .18), or rectal temperature (p = .5).
- Helmet tolerance decreased over time, with 13 dogs withdrawn for intolerance (9 CPAP, 4 control) and 6 dogs withdrawn for airway compromise (5 control, 1 CPAP).
- CPAP group maintained PaO₂/FiO₂ ratios near or above 400 mmHg, while control dogs showed consistently lower values.
- Authors suggest PaO₂/FiO₂ is a more reliable oxygenation measure than SpO₂ for evaluating postoperative lung function.
- The study found no cases of hyperthermia; temperatures normalized over time in both groups.
- Arterial catheterization failed in several dogs, limiting blood gas analysis to a subset of the population.
- Authors recommend further studies focused on BOAS patients and exploring longer-duration CPAP use to assess impact on critical outcomes like tracheostomy rates or hospitalization.
Veterinary Surgery
5
2024
Effect of continuous positive airway pressure helmet on respiratory function following laparoscopic surgery in healthy dogs
2024-5-VS-araos-2