In Lu 2025 et al., on SOP constructs, what was the **mean mediolateral bending stiffness** of SOP constructs **with** bending tees?
🔍 Key Findings
- Bending tees significantly increased mediolateral bending stiffness, but not craniocaudal stiffness, in plate-bone constructs.
- Mean mediolateral stiffness was 43.2 N/mm with tees vs. 41.1 N/mm without (p = 0.0042), though the absolute difference was small.
- No significant differences were found in craniocaudal bending stiffness between constructs with or without tees (p = 0.89).
- Plastic deformation occurred in all constructs; no screw pull-out or implant breakage was observed.
- SOP nodes may resist compressive but not tensile deformation, suggesting variable mechanical contributions depending on loading direction.
- Craniocaudal bending had greater stiffness than mediolateral due to higher area moment of inertia along the node diameter.
- Clinical relevance of added stiffness from tees remains unclear, warranting further in vivo and cyclic testing.
- This was the first study to directly test SOP constructs with/without tees over a fracture gap in multiple planes.
Veterinary and Comparative Orthopaedics and Traumatology
2
2025
Comparison of Bending Stiffness between String of Pearls Plate-Bone Substitute Constructs with and without Bending Tees in a Fracture Gap Model
2025-2-VCOT-lu-2
In Butare-Smith 2022 et al., on cerclage knot biomechanics, what was the peak load resisted by double-loop cerclage in single-load testing?
🔍 Key Findings
- Double-loop cerclage resisted the highest peak load (805 N) and maintained tension longer than twist (488 N) and single-loop (397 N) configurations.
- Double-loop cerclage sustained 500,000 cycles at 60–80% of peak load in some cases without loosening, outperforming other types.
- Twist knots loosened rapidly, often within 10 cycles even at low loads (100–390 N).
- Single-loop knots performed better than twist, with partial resistance up to 100,000 cycles at 160 N, but showed wide variability.
- All loosening occurred before wire breakage, indicating clinical failure would happen from slack, not fracture.
- Double-loop cerclage had highest initial tension (323 N) compared to single-loop (124 N) and twist (69 N).
- Fatigue limit was not identified for twist, since they all loosened early at even 20% of peak load.
- Clinical recommendation: double-loop cerclage is best for resisting repeated subfailure loading, ideal for fissure prevention or fragment stabilization.
Veterinary Surgery
2
2022
Double-loop cerclage resists greater loads for more cycles than twist and single-loop cerclage
2022-2-VS-butare-smith-4
In Folk 2025 et al., on vessel sealing device reuse, how many devices failed intraoperatively due to malfunction?
🔍 Key Findings
40 dogs underwent splenectomy using 16 bipolar vessel sealing devices (VSDs)
Devices were reused up to 4 times after handwashing and ethylene oxide (EtO) sterilization
Biologic debris was found in 100% of devices, specifically under the transection blade, even after a single use
- Mostly scant (14/16) or mild (2/16) debris
No devices or debris yielded positive aerobic culture after EtO sterilization
EtO sterilization proved microbiologically effective despite visible residue
Perioperative failure rate: 1 device (malfunctioned during first activation)
Veterinary Surgery
3
2025
Incidence of residual biologic debris and contamination of reused bipolar vessel sealing devices after ethylene oxide sterilization following splenectomy
2025-3-VS-folk-5
In Buote 2023 et al., on laparoscopic gastrectomy in cats, which staple type was ultimately used for live feline surgeries?
🔍 Key Findings
- LVSG was feasible in 9/10 feline cadavers with successful stapled gastrectomy and minimal technical complications.
- Stenosis at the incisura angularis occurred in 2/10 cadavers, associated with staple placement too close to the lesser curvature.
- Leak testing was negative in 8 cadavers and both live cats, indicating effective staple sealing.
- Mean stomach resection was ~28%, though less than human standards (~75–80%) for metabolic effects.
- Surgery was performed safely in two live feline subjects, with no intraoperative or postoperative complications over a 6-month follow-up.
- Technique refinements included orogastric tube placement and custom 3D-printed cannulas to improve staple line accuracy and avoid stenosis.
- Tri-Staple purple cartridges provided graduated compression suited for feline gastric tissue thickness (~2.5 mm).
- No need for staple line oversew in live cats; staple-only closure proved safe in this short-term study.
Veterinary Surgery
6
2023
Laparoscopic vertical sleeve gastrectomy in felines: A cadaveric feasibility study and experimental case series in two cats
2023-6-buote2-4
In Wood 2024 et al., on knot security and locking throws, which suture material showed greater holding strength?
🔍 Key Findings
- Adding a single locking throw significantly increased holding security for specific knots, including the surgeon's throw (p = .0001) and square throw (p = .0002).
- For the Miller's throw (p = .166) and strangle throw (p = .808), no significant improvement was observed with a locking throw.
- After locking throw addition, all five knots leaked at similar pressures (p = .5233), and these pressures exceeded physiologic arterial pressures.
- Surgeon's throw without a locking throw had the lowest leak pressure (62.5 ± 46.2 mm Hg), below physiologic arterial values.
- The square throw without locking also leaked below physiologic pressures (148.7 ± 109.4 mm Hg), though it outperformed the surgeon's throw.
- Miller’s and strangle throws performed significantly better than square or surgeon’s throws without locking, achieving leak pressures >200 mm Hg.
- All knots used 2-0 polyglyconate monofilament (Maxon); no comparisons across suture types or sizes were performed.
- Authors concluded that correct tensioning and locking throw addition are key to safe vascular ligation. Miller’s, strangle, or slip knots are preferred for challenging surgical fields.
Veterinary Surgery
4
2024
Influence of a single locking throw on the in vitro holding security of five friction knots using two monofilament suture materials in a canine model
2024-4-VS-wood-4
In Walter de Bruyn 2024 et al., what mechanical property was completely normalized across all working lengths after orthogonal plate application?
🔍 Key Findings Summary
- Primary 3.5-mm LCP used with short (SWL), medium (MWL), and long (LWL) working lengths
- Addition of orthogonal 2.7-mm LCP resulted in:
- Significantly higher bending stiffness for SWL, MWL, and LWL (p < 0.0001)
- Higher torsional stiffness for MWL and LWL (not for SWL)
- Significantly lower strain across all working lengths in bending (p < 0.01)
- Working length inversely related to construct stiffness and directly to plate strain
- Orthogonal plates eliminated stiffness differences across working lengths in bending
- Suggests orthogonal plates can improve implant fatigue life and allow compensation when short working lengths are unachievable
Veterinary and Comparative Orthopedics and Traumatology
4
2024
Effect of an Orthogonal Locking Plate and Primary Plate Working Length on Construct Stiffness and Plate Strain in an In vitro Fracture-Gap Model
2024-4-VCOT-walterdebruyn-4
In Trefny 2025 et al., on plate length and stiffness, what plate–bone ratio was required before significant stiffness and strain differences became apparent?
🔍 Key Findings
- 12-hole LCPs (80% plate–bone ratio) showed significantly higher construct stiffness than 6-, 8-, or 10-hole plates in both compression and tension bending.
- Strain on the plate was significantly lower in 12-hole vs 6-hole plates at all regions of interest (ROIs), especially around the fracture gap.
- No incremental increases in stiffness or decreases in strain were observed between 6-, 8-, and 10-hole plates—only when comparing to 12-hole plates.
- Bone model strain adjacent to the plate end was significantly lower for 10- and 12-hole plates vs 6-hole plates under both loading conditions.
- The threshold effect suggests biomechanical benefits only emerge beyond a plate–bone ratio of ~80%.
- Working length increased from 9.4 mm (6-hole) to 13 mm (others), potentially influencing strain/stiffness differences.
- Four-point bending was used, as it replicates the most biomechanically relevant force on plated long bones.
- Clinical implication: Longer plates may reduce plate strain and peri-implant bone strain, potentially lowering risk of fatigue failure or stress risers.
Veterinary and Comparative Orthopaedics and Traumatology
2
2025
Effect of Plate Length on Construct Stiffness and Strain in a Synthetic Short-Fragment Fracture Gap Model Stabilized with a 3.5-mm Locking Compression Plate
2025-2-VCOT-trefny-5
In Hawker 2025 et al., on locking head inserts, what was the impact of LHI on plate strain in a fracture-gap model?
🔍 Key Findings
- Adding Locking Head Inserts (LHI) to a 3.5-mm LCP had no effect on plate strain, stiffness, or deformation in an open fracture gap model.
- Peak strain consistently occurred at the Combi-hole over the fracture gap, with values up to ~1837 µε.
- No significant difference in strain was found across configurations with 0, 3, or 9 LHI (p = 0.847).
- Construct stiffness and compressive displacement also remained unchanged regardless of LHI count (p = 0.311 and 0.069 respectively).
- Study contradicted the hypothesis that LHI would reduce strain and increase stiffness under biologic loading.
- Combi-hole design may limit the efficacy of LHI, as LHI only fill the locking portion, not the compression side where strain peaks.
- Implant fatigue risk remains highest over unfilled screw holes, especially over fracture sites—confirming previous failure patterns.
- Surgeons should consider alternative methods to reduce strain when facing high implant load scenarios.
Veterinary and Comparative Orthopaedics and Traumatology
4
2025
The Effect of Locking Head Inserts on the Biomechanical Properties of a 3.5-mm Broad Locking Compression Plate When Used in an Open Fracture-Gap Model
2025-4-VCOT-hawker-1
In Trefny 2025 et al., on locking plate biomechanics, which configuration showed higher construct stiffness in compression bending?
🔍 Key Findings
- Short working length constructs had significantly higher stiffness and lower strain than long constructs in compression bending (p = 0.0172).
- In tension bending, short constructs also had higher precontact stiffness and lower strain, but this reversed after transcortical contact (~150 N).
- Transcortical contact increased stiffness only in long constructs, producing a bilinear load-displacement curve.
- Postcontact stiffness was higher in long constructs, but this may not reflect clinical benefit due to risks of high interfragmentary strain.
- Short working length reduced strain at multiple ROIs under both loading conditions, including over fracture gap (Tables 1–3).
- Increased working length promoted stress concentration and deformation, especially in compression bending.
- In vitro benefits of long constructs (via contact stability) may not translate to healing, as repetitive loading could increase plate strain and bone resorption.
- Plate strain was effectively mapped using 3D digital image correlation, confirming regional strain differences between configurations.
Veterinary and Comparative Orthopaedics and Traumatology
3
2025
Effect of Plate Screw Configuration on Construct Stiffness and Plate Strain in a Synthetic Short Fragment Small Gap Fracture Model Stabilized with a 12-Hole 3.5-mm Locking Compression Plate
2025-3-VCOT-trefny-1
In Hawker 2025 et al., on locking head inserts, what effect did LHI have on axial stiffness and displacement?
🔍 Key Findings
- Adding Locking Head Inserts (LHI) to a 3.5-mm LCP had no effect on plate strain, stiffness, or deformation in an open fracture gap model.
- Peak strain consistently occurred at the Combi-hole over the fracture gap, with values up to ~1837 µε.
- No significant difference in strain was found across configurations with 0, 3, or 9 LHI (p = 0.847).
- Construct stiffness and compressive displacement also remained unchanged regardless of LHI count (p = 0.311 and 0.069 respectively).
- Study contradicted the hypothesis that LHI would reduce strain and increase stiffness under biologic loading.
- Combi-hole design may limit the efficacy of LHI, as LHI only fill the locking portion, not the compression side where strain peaks.
- Implant fatigue risk remains highest over unfilled screw holes, especially over fracture sites—confirming previous failure patterns.
- Surgeons should consider alternative methods to reduce strain when facing high implant load scenarios.
Veterinary and Comparative Orthopaedics and Traumatology
4
2025
The Effect of Locking Head Inserts on the Biomechanical Properties of a 3.5-mm Broad Locking Compression Plate When Used in an Open Fracture-Gap Model
2025-4-VCOT-hawker-3