Quiz Question
In Deprey 2022 et al., on gap fracture implants, which construct demonstrated higher torque to failure during torsional testing?
🔍 Key Findings
- NAS-ILN had significantly greater stiffness in both axial compression and 4-point bending compared to LCP constructs.
- Ultimate load to failure was significantly higher for NAS-ILN in compression (804 N vs 328 N) and bending (25.7 Nm vs 16.3 Nm).
- Torsional stiffness and angular deformation were similar, but NAS-ILN resisted higher torque to failure than LCP (22.5 Nm vs 19.1 Nm).
- No slack was observed with the NAS-ILN construct, unlike older nail designs.
- Failure modes differed: LCPs failed via plate bending; NAS-ILNs failed at the implant or bone near screw holes.
- Titanium alloy and curved design of NAS-ILN provides better anatomic fit and more uniform stress distribution.
- A third, perpendicular locking hole in NAS-ILN may enhance torsional stability but was not utilized in this study.
- The curved, angle-stable design of NAS-ILN is a novel advancement in veterinary orthopedics.
Veterinary Surgery
8
2022
Mechanical evaluation of a novel angle‐stable interlocking nail in a gap fracture model
2022-8-VS-deprey-3
In Butare-Smith 2022 et al., on cerclage knot biomechanics, which knot type demonstrated the highest resistance to cyclic loading before loosening?
🔍 Key Findings
- Double-loop cerclage resisted the highest peak load (805 N) and maintained tension longer than twist (488 N) and single-loop (397 N) configurations.
- Double-loop cerclage sustained 500,000 cycles at 60–80% of peak load in some cases without loosening, outperforming other types.
- Twist knots loosened rapidly, often within 10 cycles even at low loads (100–390 N).
- Single-loop knots performed better than twist, with partial resistance up to 100,000 cycles at 160 N, but showed wide variability.
- All loosening occurred before wire breakage, indicating clinical failure would happen from slack, not fracture.
- Double-loop cerclage had highest initial tension (323 N) compared to single-loop (124 N) and twist (69 N).
- Fatigue limit was not identified for twist, since they all loosened early at even 20% of peak load.
- Clinical recommendation: double-loop cerclage is best for resisting repeated subfailure loading, ideal for fissure prevention or fragment stabilization.
Veterinary Surgery
2
2022
Double-loop cerclage resists greater loads for more cycles than twist and single-loop cerclage
2022-2-VS-butare-smith-1
In Fitzpatrick 2024 et al., on ESF for pelvic fractures in cats, what percentage of cases showed implant loosening on follow-up radiographs?
🔍 Key Findings
- External skeletal fixation (ESF) was successfully applied to a variety of pelvic fracture types in cats, including sacroiliac luxations and ilial body fractures.
- All fractures achieved radiographic union within 9 weeks, even in comminuted or complex configurations.
- No intraoperative or long-term complications were reported during the study period.
- Implant loosening was observed radiographically in 13% of cases, with 8% of pins found to be loose at frame removal.
- ESF enabled indirect fracture reduction using components as handles, with a limited open approach minimizing soft tissue disruption.
- No iatrogenic neurological deficits were observed, supporting safe pin placement near neurovascular structures, although some cats presented with pre-existing neurologic signs.
- No cases required revision surgery, and all cats underwent stabilization solely with ESF as per study inclusion criteria.
- Postoperative hospitalization ranged from 2 to 5 days, though no comparison to other fixation types was evaluated.
Veterinary Surgery
7
2024
External skeletal fixation for the treatment of pelvic fractures in cats
2024-7-VS-fitzpatrick-2
In McCarthy 2022 et al., on 3D drill guide accuracy, what was the advantage of 3D-GDT in terms of drill end point deviation?
🔍 Key Findings
- 3D-printed drill guides (3D-GDT) significantly reduced craniocaudal and dorsoventral drilling angle deviation compared to free-hand drilling technique (FHDT), with statistical significance (p < .0001 and p = .01 respectively).
- No sacral corridor breaches occurred with 3D-GDT, whereas FHDT resulted in 20% drill exit incidences (3/15 cases).
- 3D-GDT had lower deviation from optimal drill trajectory at end points in all axes: craniocaudal (1.84 ± 1.6 mm vs. 4.18 ± 2.4 mm), dorsoventral (1.11 ± 1.0 mm vs. 2.4 ± 1.5 mm), and 3D linear (2.47 ± 1.4 mm vs. 5.35 ± 2.2 mm), all statistically significant.
- Strong correlation (r = 0.77) between CT and 3D software measurements validated method reliability.
- 3D guide trajectories showed consistent proximity to the optimal trajectory, especially at drill endpoints, indicating reduced variability in execution.
- 3D-GDT was developed using open-source software and inexpensive materials, supporting future clinical application.
- Major errors in FHDT occurred primarily at the drill endpoint, underscoring the challenge of maintaining optimal angulation during free-hand drilling.
- The sacral corridor was recharacterized as pyramidal (not quadrilateral), with implications for safer implant placement.
Veterinary Surgery
1
2022
Accuracy of a drilling with a custom 3D printed guide or free‐hand technique in canine experimental sacroiliac luxations
2022-1-VS-mccarthy-3
In Mullen 2023 et al., on microvascular perfusion, why was the green TA (4.8 mm) cartridge selected for stapled enterectomies?
🔍 Key Findings
- Microvascular density at the site of foreign body obstruction was significantly lower in obstructed dogs compared to healthy controls.
- PBR (perfused boundary region) was paradoxically lower in obstructed dogs, indicating that PBR may not reliably correlate with tissue viability in this context.
- No significant difference in microvascular perfusion between subjectively viable vs nonviable intestines, suggesting subjective assessment is unreliable.
- No difference in perfusion parameters between handsewn and stapled enterectomies, indicating both techniques preserve vascular health equally.
- Stapled enterectomy using green TA staples is safe and did not impair intestinal perfusion at the anastomosis site.
- SDF videomicroscopy is feasible intraoperatively and can differentiate healthy from diseased intestine in dogs.
- Surgeons' subjective evaluations led to potentially unnecessary enterectomies, as microvascular parameters did not differ.
- No cases of dehiscence or mortality, suggesting both surgical approaches are effective when applied properly.
Veterinary Surgery
4
2023
A quantitative evaluation of the effect of foreign body obstruction and enterectomy technique on canine small intestinal microvascular health
2023-4-VS-mullen-5
In Rodiño Tilve 2022 et al., on feline THR outcomes, what femoral implant size was most commonly used?
🔍 Key Findings
From “Long-term follow up of 44 cats undergoing total hip replacement” by Rodiño Tilve et al.
- Slipped capital femoral epiphysis (SCFE) was the most common surgical indication (61%, 34/56 hips), primarily affecting young neutered male cats.
- All THRs used cemented micro/nano BioMedtrix implants; most common femoral stem was size #3, and most common acetabular cup was 12 mm.
- Postoperative complication rate was 19.6% (11/56) with 9 major complications (luxation most common), and no intraoperative complications reported.
- All luxations occurred in hips implanted with femoral neck +0 mm length implants.
- Second luxations were more common when revision used same implant size; use of larger implants reduced reluxation rates.
- FMPI-sf score improved significantly from median 2.111 pre-op to 0.111 post-op (P < .001), indicating reduced pain and improved function.
- Very high owner satisfaction: 91% (30/33) reported outcome as "very good."
- No significant associations found between complications and variables like weight, sex, implant size, or surgical indication.
Veterinary Surgery
5
2022
Long‐term follow up of 44 cats undergoing total hip replacement: Cases from a feline hip registry (2010–2020)
2022-5-VS-rodino-2
In Huels 2025 et al., on second-generation screw cup THA, what was the observed rate of late aseptic loosening?
🔍 Key Findings
- Total complication rate was 16.7%, with 5/30 hips experiencing major complications, mostly related to the femoral component.
- Cup-associated complications were rare (3.3%), with only one case of acetabular cup luxation attributed to surgical technique rather than implant failure.
- No cases of late aseptic loosening were observed during a median follow-up of 17.5 months.
- Implant stability was attributed to the SCSL's porous, trabecular titanium surface, enhancing osseointegration.
- Three femoral stem fractures occurred in a single dog, leading to implant removal; material testing was not performed.
- Most complications were femoral in origin (6/7), not acetabular, suggesting improved performance of the SCSL.
- Explantation rate was 13% (4/30), but some removals were due to owner preference against revision.
- Subjective functional outcome was full recovery in 26/30 hips, including one with successful revision of stem subsidence.
Veterinary and Comparative Orthopaedics and Traumatology
2
2025
Complications and Long-Term Outcome in 30 Canine Total Hip Arthroplasties Using a Second-Generation Selective Laser Melted Screw Cup
2025-2-VCOT-huels-1
In Scortea 2025 et al., on sacroiliac fixation accuracy, which statement regarding sacral bone purchase is correct?
🔍 Key Findings
- 3D-printed drill guide technique (3D-DGT) resulted in fewer suboptimal screw placements than minimally invasive osteosynthesis (MIO) (7.14% vs 42.85%), though not statistically significant.
- Entry point translation (EPT) in the dorsoventral direction was significantly lower with 3D-DGT compared with MIO (p = .009).
- Maximum angular screw deviation (MASD) did not differ significantly between 3D-DGT and MIO in dorsal or transverse planes.
- Ventral cortical breach was the most common error with both techniques, more frequent with MIO.
- Achievement of >60% sacral bone purchase was more consistent with 3D-DGT (92.9%) than with MIO (64.3%).
- Modified Gras grade distribution was similar between techniques, with most screws graded as secure (“a” or “b”).
- 3D-DGT required substantially longer preoperative planning time than MIO (median 34 vs 8.5 minutes).
- Both techniques demonstrated overall acceptable accuracy, emphasizing the importance of CT-based planning and assessment.
Veterinary and Comparative Orthopaedics and Traumatology
6
2025
Comparative analysis of 3D-printed drill guides and minimally invasive osteosynthesis in feline sacroiliac luxation: A cadaveric study
2025-6-VCOT-scortea-4
In Trefny 2025 et al., on locking plate biomechanics, which configuration showed higher construct stiffness in compression bending?
🔍 Key Findings
- Short working length constructs had significantly higher stiffness and lower strain than long constructs in compression bending (p = 0.0172).
- In tension bending, short constructs also had higher precontact stiffness and lower strain, but this reversed after transcortical contact (~150 N).
- Transcortical contact increased stiffness only in long constructs, producing a bilinear load-displacement curve.
- Postcontact stiffness was higher in long constructs, but this may not reflect clinical benefit due to risks of high interfragmentary strain.
- Short working length reduced strain at multiple ROIs under both loading conditions, including over fracture gap (Tables 1–3).
- Increased working length promoted stress concentration and deformation, especially in compression bending.
- In vitro benefits of long constructs (via contact stability) may not translate to healing, as repetitive loading could increase plate strain and bone resorption.
- Plate strain was effectively mapped using 3D digital image correlation, confirming regional strain differences between configurations.
Veterinary and Comparative Orthopaedics and Traumatology
3
2025
Effect of Plate Screw Configuration on Construct Stiffness and Plate Strain in a Synthetic Short Fragment Small Gap Fracture Model Stabilized with a 12-Hole 3.5-mm Locking Compression Plate
2025-3-VCOT-trefny-1
In Wilson 2025 et al., on acetabular measurement accuracy, what was the main drawback of femoral head circle methods?
🔍 Key Findings
Study population: 73 hips from 60 dogs undergoing cementless THR.
Methods evaluated:
- ACVD/ACOLL (acetabular circle on VD or OLL view)
- ALVD/ALOLL (acetabular line)
- FHCVD/FHCOLL/FHCCCHB (femoral head circle)
- Intraobserver repeatability and interobserver consistency were excellent for ACVD and ACOLL.
- FHC methods consistently underestimated actual cup size by 2.4–3.6 mm.
- AC and AL methods had low bias (±0.5 mm) and better predictive value.
- OA severity negatively affected the accuracy of all measurements (p < .05).
- Highest predictive accuracy was ~49% using ACVD with rounding down protocol.
Veterinary Surgery
1
2025
Evaluation of three acetabular measurement methods for total hip replacement in dogs
2025-1-VS-wilson-2
Quiz Results
You answered 7 out of 10 questions correctly
Key Findings