In Chen 2024 et al., on pressure-measurement tools, how much does the risk of poor outcome increase per 1 mm Hg rise in portal pressure during complete PSS occlusion?
🔍 Key Findings
- WMg (water manometer with gauge) was the most accurate and precise pressure measurement device.
- APT (arterial pressure transducer) was less accurate than WMg but still precise; it differed significantly from the gold standard (WMr).
- CCT (Compass CT) was the least accurate and precise and differed significantly from the set pressure.
- Mean differences from set pressure were smallest for WMg (−0.020 cm H2O), moderate for APT (−0.390 cm H2O), and largest for CCT (−1.267 cm H2O).
- All devices showed excellent interobserver (ICC = 1.000) and intraobserver agreement (ICC range 0.985–0.998).
- Even though the CCT performed least well, all devices had mean errors ≤1.3 cm H2O, indicating potential clinical utility.
- WMg or WMr should be preferred in surgical settings due to superior accuracy and precision.
- Measurement precision is more critical than accuracy during PSS surgery, as a 1 mm Hg (~1.3 cm H2O) increase in portal pressure raises odds of poor outcomes by 9%.
Veterinary Surgery
4
2024
Accuracy, precision, and interobserver and intraobserver agreements related to pressure-measurement devices
2024-4-VS-chen-5
In Vodnarek 2024 et al., on intraobserver performance, which observer achieved **excellent reliability** for both methods?
🔍 Key Findings
- Study population: 36 brachycephalic dogs (20 French bulldogs, 16 pugs).
- Objective: Compare intra- and interobserver reliability for fluoroscopic measurement of nasopharyngeal collapse using two methods:
- Functional method
- Anatomically adjusted method
- Key measurements: Minimum (LMin), maximum (LMax) dorsoventral height, and dynamic change ratio (ΔL).
- Outcomes:
- Intraobserver agreement for ΔL was higher with the functional method (ICC 0.751 vs. 0.576).
- Observer 1 (radiologist) showed excellent repeatability (>0.9 ICC).
- Agreement for grading collapse was only moderate (κ ~0.49–0.53), worse than ΔL-based agreement.
- ΔL ≥ 0.5 to <1 = partial collapse; ΔL = 1 = complete collapse.
Veterinary Surgery
1
2024
Reliability of fluoroscopic examination of nasopharyngeal dorsoventral dimension change in pugs and French bulldogs
2024-1-VS-vodnarek-3
In Banks 2023 et al., on TECA-LBO in brachycephalic dogs, what was the overall perioperative complication rate and how did it compare between groups?
🔍 Key Findings
- Extreme brachycephalic breeds (EBBs) presented more acutely and at younger ages, most often with neurological signs compared to other breeds.
- Preoperative signs such as facial nerve paresis, vestibular syndrome, and Horner’s syndrome were significantly more common in EBBs.
- EBBs showed more severe imaging findings, including higher rates of otitis interna (46.3% vs. 8.5%) and brainstem changes (17.5% vs. 3%).
- Intraoperative complications were more frequent in EBBs (11.1% vs. 5.3%), but this difference did not reach statistical significance (p = .078).
- Perioperative complication rates did not differ significantly between EBBs and other breeds (23.5% vs. 29.3%).
- Surgical time was significantly longer in EBBs (median 115 vs. 95 minutes; p = .011).
- MRI or combined CT/MRI were more frequently used in EBBs, likely due to the higher prevalence of neurological signs.
- Despite anatomical challenges, complication rates in EBBs were comparable, supporting TECA-LBO safety in these breeds.
Veterinary Surgery
5
2023
Influence of extreme brachycephalic conformation on perioperative complications associated with total ear canal ablation and lateral bulla osteotomy in 242 dogs (2010–2020)
2023-5-VS-banks-5
In Healy 2025 et al., on incidental PBBs, what was the prevalence of PBBs in the studied dog population?
🔍 Key Findings
Population: 2,178 canine CTs reviewed retrospectively.
Prevalence: Incidental PBBs found in 1.37% (30/2178).
Outcome: None of the dogs with incidental PBBs developed clinical spontaneous pneumothorax (SP) over a median follow-up of 1255 days.
Significant Associations:
- Age: Dogs with PBBs were significantly older (median 10.5 yrs vs. 8.2 yrs, p = .001).
- CT indication: PBBs more likely during neoplastic staging (p = .006).
PBB Characteristics:
- Total = 60 PBBs (median 1/dog; range 1–7).
- Location: 35% in left caudal, 31.6% right caudal, only 13.3% in right cranial lobe.
- Size-based: 25 bullae (>10 mm), 35 blebs (≤10 mm).
Conclusion: Prophylactic resection of incidental PBBs not justified given no observed SP risk in this population.
Veterinary Surgery
1
2025
Significance of incidentally identified bullae and blebs on thoracic computed tomography and prevalence of subsequent pneumothorax in dogs
2025-1-VS-healy-1
In Healy 2025 et al., on incidental PBBs, which lung lobe was most frequently affected?
🔍 Key Findings
Population: 2,178 canine CTs reviewed retrospectively.
Prevalence: Incidental PBBs found in 1.37% (30/2178).
Outcome: None of the dogs with incidental PBBs developed clinical spontaneous pneumothorax (SP) over a median follow-up of 1255 days.
Significant Associations:
- Age: Dogs with PBBs were significantly older (median 10.5 yrs vs. 8.2 yrs, p = .001).
- CT indication: PBBs more likely during neoplastic staging (p = .006).
PBB Characteristics:
- Total = 60 PBBs (median 1/dog; range 1–7).
- Location: 35% in left caudal, 31.6% right caudal, only 13.3% in right cranial lobe.
- Size-based: 25 bullae (>10 mm), 35 blebs (≤10 mm).
Conclusion: Prophylactic resection of incidental PBBs not justified given no observed SP risk in this population.
Veterinary Surgery
1
2025
Significance of incidentally identified bullae and blebs on thoracic computed tomography and prevalence of subsequent pneumothorax in dogs
2025-1-VS-healy-5
In Miller 2024 et al., on staple vs. hand-sewn feline GI techniques, what did the authors conclude about skin staple enterotomy (SSE) for use in live cats?
🔍 Key Findings
- Skin staple anastomosis (SSA) had comparable leak pressures to hand-sewn anastomosis (HSA) but required half the time to complete.
- Skin staple enterotomy (SSE) had significantly lower leak pressures than hand-sewn enterotomy (HSE) and failed in 12/20 constructs during pressure testing.
- HSE constructs took 8× longer to complete than SSE, but had much higher intraluminal pressure tolerance.
- All SSE constructs leaked from the center, with 35% leaking immediately and 60% showing catastrophic failure.
- SSA leakage occurred at the center in 40% of constructs, likely due to a learning curve in early samples.
- All constructs had higher pressures than normal physiologic intestinal pressure (4.0 mmHg ±2.0), except some SSEs with immediate leaks.
- Authors recommend SSA as a viable alternative with appropriate training but do not recommend SSE using the tested technique in live cats.
- Staple size and placement technique are key factors; smaller or more precisely placed staples may reduce leak risk.
Veterinary Surgery
4
2024
Performance time and leak pressure of hand-sewn and skin staple intestinal anastomoses and enterotomies in cadaveric cats
2024-4-VS-miller-5
In Sisk 2024 et al., what conclusion did the authors draw regarding reamed versus unreamed IMN in dogs?
🔍 Key Findings Summary
- IMN provides relative stability, resists bending/torsion due to central axis alignment
- Larger diameter nails = exponentially greater stiffness (∝ D⁴)
- Trade-off: Larger interlocking holes weaken fatigue strength of the nail
- Reaming increases contact/stability but has pros/cons:
- Improves outcomes in closed fractures
- May reduce endosteal blood flow in thin-walled bones (e.g., cats)
- Design advances:
- Angle-stable IMN reduce rotational slack
- Expandable nails simplify insertion but may compromise removal or compressive load resistance
- Precontoured nails match bone curvature but lack consistent clinical superiority
- Material debates continue (e.g., titanium vs. stainless steel vs. magnesium)
Veterinary and Comparative Orthopedics and Traumatology
6
2024
Biomechanical Principles of Intramedullary Nails in Veterinary and Human Medicine
2024-6-VCOT-sisk-5
In Scott 2025 et al., on acetabular cup revision, what was the functional outcome in dogs that completed follow-up?
🔍 Key Findings
Population: 9 dogs underwent revision of osteointegrated acetabular cups after total hip arthroplasty (THA)
Revision Indications:
- 7 luxations (5 ventral, 2 craniodorsal)
- 1 femoral stem fracture
- 1 aseptic stem loosening
Implants:
- 8 BFX cups, 1 Helica; all revised to BFX
- 7/9 required a larger cup than original
Cup removal: Required sectioning with a high-speed burr and modular osteotome; removal fragments extracted
Complications:
- 1 recurrent luxation
- 1 low-grade infection with possible metallic debris-associated osteolysis
- 2 femoral fissures managed intraoperatively
Outcomes:
- Good to excellent function in 6/6 dogs available at median 621 days
- Minimal complications with success in re-osteointegration of new cup
Clinical takeaway: Revision of stable, ingrown cups is feasible and offers an alternative to pelvic osteotomies; typically requires upsizing
Veterinary Surgery
3
2025
Revision of osteointegrated acetabular cup prostheses in nine dogs
2025-3-VS-scott-5
In Scharpf 2024 et al., what type of analysis was used to assess limb loading recovery over time?
🔍 Key Findings Summary
- Subtotal coronoidectomy improved vertical and propulsive forces, but braking forces remained subnormal at 26 weeks.
- No significant benefit was seen from ACP vs placebo at any timepoint across all force parameters or lameness scores.
- Force plate analysis was more sensitive than visual lameness scoring.
- Braking force (%FY+) was best at detecting persistent lameness, and SI < 0.9 persisted in most dogs at 26 weeks.
- Outcome less favorable than historically reported — challenges status of subtotal coronoidectomy as “gold standard” for MCD.
Veterinary and Comparative Orthopedics and Traumatology
2
2024
Assessment of Arthroscopic Subtotal Coronoidectomy in Treating Medial Coronoid Disease and Effect of Concurrent Autologous Conditioned Plasma in Dogs Using Force Plate Analysis
2024-2-VCOT-scharpf-4
In Smith 2025 et al., on ergonomic injury risk, which of the following was NOT significantly associated with musculoskeletal disorder prevalence?
🔍 Key Findings
140 laparoscopic surgeons surveyed; 37% reported at least one musculoskeletal disorder.
Women had significantly higher odds of reporting WRMD (OR = 2.59, p = .011).
Smaller glove size significantly associated with WRMD (p = .001), shoulder tendonitis (p = .01), and neck strain (p = .001).
Most common injuries: Neck strain (35%), shoulder tendonitis (31%).
WRMD was associated with greater difficulty using:
- Rotating cup biopsy forceps (p < .001)
- Vessel sealing device and endo stapler (especially in those with shoulder injuries)