🔍 Key Findings
- 12-hole LCPs (80% plate–bone ratio) showed significantly higher construct stiffness than 6-, 8-, or 10-hole plates in both compression and tension bending.
- Strain on the plate was significantly lower in 12-hole vs 6-hole plates at all regions of interest (ROIs), especially around the fracture gap.
- No incremental increases in stiffness or decreases in strain were observed between 6-, 8-, and 10-hole plates—only when comparing to 12-hole plates.
- Bone model strain adjacent to the plate end was significantly lower for 10- and 12-hole plates vs 6-hole plates under both loading conditions.
- The threshold effect suggests biomechanical benefits only emerge beyond a plate–bone ratio of ~80%.
- Working length increased from 9.4 mm (6-hole) to 13 mm (others), potentially influencing strain/stiffness differences.
- Four-point bending was used, as it replicates the most biomechanically relevant force on plated long bones.
- Clinical implication: Longer plates may reduce plate strain and peri-implant bone strain, potentially lowering risk of fatigue failure or stress risers.
Simini Surgery Review Podcast
🔍 Key Findings
- 12-hole LCPs (80% plate–bone ratio) showed significantly higher construct stiffness than 6-, 8-, or 10-hole plates in both compression and tension bending.
- Strain on the plate was significantly lower in 12-hole vs 6-hole plates at all regions of interest (ROIs), especially around the fracture gap.
- No incremental increases in stiffness or decreases in strain were observed between 6-, 8-, and 10-hole plates—only when comparing to 12-hole plates.
- Bone model strain adjacent to the plate end was significantly lower for 10- and 12-hole plates vs 6-hole plates under both loading conditions.
- The threshold effect suggests biomechanical benefits only emerge beyond a plate–bone ratio of ~80%.
- Working length increased from 9.4 mm (6-hole) to 13 mm (others), potentially influencing strain/stiffness differences.
- Four-point bending was used, as it replicates the most biomechanically relevant force on plated long bones.
- Clinical implication: Longer plates may reduce plate strain and peri-implant bone strain, potentially lowering risk of fatigue failure or stress risers.
Simini Surgery Review Podcast
Know What Matters in the Literature - and Why
We distill peer-reviewed surgical studies into clinically relevant summaries and
exam-style questions, so you can make informed decisions with confidence.
Free Access. No Spam. Just Smarter Surgical Learning
Multiple Choice Questions on this study
Access the full library of surgical summaries and exam-style questions.