🔍 Key Findings
- 3D-printed drill guides (3D-GDT) significantly reduced craniocaudal and dorsoventral drilling angle deviation compared to free-hand drilling technique (FHDT), with statistical significance (p < .0001 and p = .01 respectively).
- No sacral corridor breaches occurred with 3D-GDT, whereas FHDT resulted in 20% drill exit incidences (3/15 cases).
- 3D-GDT had lower deviation from optimal drill trajectory at end points in all axes: craniocaudal (1.84 ± 1.6 mm vs. 4.18 ± 2.4 mm), dorsoventral (1.11 ± 1.0 mm vs. 2.4 ± 1.5 mm), and 3D linear (2.47 ± 1.4 mm vs. 5.35 ± 2.2 mm), all statistically significant.
- Strong correlation (r = 0.77) between CT and 3D software measurements validated method reliability.
- 3D guide trajectories showed consistent proximity to the optimal trajectory, especially at drill endpoints, indicating reduced variability in execution.
- 3D-GDT was developed using open-source software and inexpensive materials, supporting future clinical application.
- Major errors in FHDT occurred primarily at the drill endpoint, underscoring the challenge of maintaining optimal angulation during free-hand drilling.
- The sacral corridor was recharacterized as pyramidal (not quadrilateral), with implications for safer implant placement.
Simini Surgery Review Podcast
🔍 Key Findings
- 3D-printed drill guides (3D-GDT) significantly reduced craniocaudal and dorsoventral drilling angle deviation compared to free-hand drilling technique (FHDT), with statistical significance (p < .0001 and p = .01 respectively).
- No sacral corridor breaches occurred with 3D-GDT, whereas FHDT resulted in 20% drill exit incidences (3/15 cases).
- 3D-GDT had lower deviation from optimal drill trajectory at end points in all axes: craniocaudal (1.84 ± 1.6 mm vs. 4.18 ± 2.4 mm), dorsoventral (1.11 ± 1.0 mm vs. 2.4 ± 1.5 mm), and 3D linear (2.47 ± 1.4 mm vs. 5.35 ± 2.2 mm), all statistically significant.
- Strong correlation (r = 0.77) between CT and 3D software measurements validated method reliability.
- 3D guide trajectories showed consistent proximity to the optimal trajectory, especially at drill endpoints, indicating reduced variability in execution.
- 3D-GDT was developed using open-source software and inexpensive materials, supporting future clinical application.
- Major errors in FHDT occurred primarily at the drill endpoint, underscoring the challenge of maintaining optimal angulation during free-hand drilling.
- The sacral corridor was recharacterized as pyramidal (not quadrilateral), with implications for safer implant placement.
Simini Surgery Review Podcast
Know What Matters in the Literature - and Why
We distill peer-reviewed surgical studies into clinically relevant summaries and
exam-style questions, so you can make informed decisions with confidence.
Free Access. No Spam. Just Smarter Surgical Learning
Multiple Choice Questions on this study
Access the full library of surgical summaries and exam-style questions.